Thermocouple Data Logger TC-08 - Input Channel Isolation

Post any questions you may have about our current range of USB data loggers
Post Reply
lightsphere
Newbie
Posts: 0
Joined: Thu May 07, 2015 2:03 pm

Thermocouple Data Logger TC-08 - Input Channel Isolation

Post by lightsphere » Thu May 07, 2015 2:11 pm

As a cannot find the answer in the products data sheets. Can you tell me if the 8 input channels of the TC-08 are isolated against each other?

We need to sample temperatures from a metal pipe where we apply 8 thermocouples type-k to the same pipe which then is heated up. Most dataloggers we found so far, have no isolation of their input channels against each other which leads to interference of the temperature measurements if there is a common ground.

Thanks for any advice.

Gerry
PICO STAFF
PICO STAFF
Posts: 594
Joined: Mon Aug 11, 2014 11:14 am

Re: Thermocouple Data Logger TC-08 - Input Channel Isolation

Post by Gerry » Tue May 12, 2015 4:27 pm

Hi Lightsphere,

The TC-08 channels are not isolated from each other they have weak pull down resistors, and capacitors that are tied to ground.

Best regards,

Gerry
Gerry
Technical Specialist

alan
Site Admin
Site Admin
Posts: 190
Joined: Thu May 25, 2006 8:02 am

Re: Thermocouple Data Logger TC-08 - Input Channel Isolation

Post by alan » Tue May 12, 2015 5:47 pm

Hi,

Just to add to this, the inputs are differential with a 7.5V common mode voltage. To put this another way as long as there is no more than ±7.5V between the thermocouples you will be OK. As you mention they are connected to the same pipe it seems unlikely there would be more than a few mV voltage drop along the pipe so it should be OK.

Post Reply