TC-08 - Voltage measuring

Post any questions you may have about our current range of USB data loggers

TC-08 - Voltage measuring

Postby battywoo » Fri Mar 20, 2015 6:35 pm

We currently use TC-08 for temperature and voltage measuring. Channel 1-6 are for thermocouples inputs while channel 7-8 are for voltage measuring (0-5V). We got two TC-08 Single-Channel Terminal Boards to convert the voltage. (switch set to OFF).

The problem is that when we try to measure the voltages from a circuit, the readings of all channels get crazy. Seems that the introduction of voltages interfere the whole system. The circuit is powered by a power supply and the TC-08 is connected to a laptop with power plugged. They are connected to the same power outlet.

BTW, if we try to measure the voltage of a AA battery, the readings seem good.
Could you please provide us some ideas to solve this problem? Thanks
battywoo
Newbie
 
Posts: 0
Joined: Fri Mar 20, 2015 6:18 pm

Re: TC-08 - Voltage measuring

Postby Martyn » Tue Mar 24, 2015 2:40 pm

Try with the laptop running on it's battery, and if this doesn't fix things then try to make sure that there is a common ground between your circuit and the laptop, plugging into the same power outlet may not be sufficient.

What are the thermocouples connected to and measuring ?
Martyn
Technical Specialist
Martyn
Site Admin
Site Admin
 
Posts: 2355
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: TC-08 - Voltage measuring

Postby battywoo » Tue Mar 24, 2015 3:18 pm

Thank you for the reply. Laptop running on battery didn't fix that.

The thermocouples are grounded K-type thermocouples to measure the temperatures in a burner. Some of them touch the metal surface. Some are not.

I will try to connect the common ground of the circuit to the burner's metal surface to see if that works.
battywoo
Newbie
 
Posts: 0
Joined: Fri Mar 20, 2015 6:18 pm

Re: TC-08 - Voltage measuring

Postby fredrik » Fri Oct 30, 2015 11:58 am

Battywoo: did it fix your problem? I have the same issue. If i put in the single channel terminal board and measure voltage (0-5V) the thermocouple readings go crazy. If I then connect the terminal board to ground the thermocouples are good but the voltage measurement is wrong (more than if not grounded). Can also this be a problem of having different earthing locations for the computer and the terminal boards?

Previously I solved this by having only one terminal board per TC08 logger, but it is an expensive solution.
fredrik
Newbie
 
Posts: 0
Joined: Fri Oct 30, 2015 11:49 am

Re: TC-08 - Voltage measuring

Postby Hitesh » Wed Nov 04, 2015 10:31 am

Hi fredrik,

Try ensuring that the computer and terminal boards are using a common ground reference.

Are the thermocouples touching any ceramic surfaces at high temperatures or metal surfaces?

Regards,
Hitesh

Technical Specialist
Pico Technology
Hitesh
Site Admin
Site Admin
 
Posts: 2063
Joined: Tue May 31, 2011 3:43 pm
Location: St. Neots, Cambridgeshire

Re: TC-08 - Voltage measuring

Postby fredrik » Thu Nov 12, 2015 7:46 am

Hi Hatesh. Thanks for answering

I have tried that but it doesn't help. Below is the result of some of the measurements I did when trying to fix this. I will look into this more tomorrow, and it would be great with some advice.

With no earthing I get the following readings when both terminal boards are insterted in the logger:
condensing: 11,9 mV
evaporating: 4,76 mV

with only the condensing inserted:
condensing: 10,5 mV

with only evaporating inserted:
evaporating: 4,22 mV

only condensing with earting of the terminal
condensing: 24mV

Regards
Fredrik
fredrik
Newbie
 
Posts: 0
Joined: Fri Oct 30, 2015 11:49 am

Re: TC-08 - Voltage measuring

Postby Hitesh » Fri Nov 13, 2015 11:37 am

Hi Fredrik,

Please ensure that the voltages being input to the TC-08 after the scale-down by the resistors does not exceed the Common Mode voltage range (+/- 7.5V).

If you could provide a diagram or photograph showing the connections to the terminal board and where they connect to, that might be helpful.

Regards,
Hitesh

Technical Specialist
Pico Technology
Hitesh
Site Admin
Site Admin
 
Posts: 2063
Joined: Tue May 31, 2011 3:43 pm
Location: St. Neots, Cambridgeshire

Re: TC-08 - Voltage measuring

Postby fredrik » Tue Nov 17, 2015 1:44 pm

Hi again

Problem solved. It was correct to connect the terminal boards to earth. We had done it wrong previously and I got confused.
fredrik
Newbie
 
Posts: 0
Joined: Fri Oct 30, 2015 11:49 am

Re: TC-08 - Voltage measuring

Postby Hitesh » Tue Nov 17, 2015 5:11 pm

Hi Fredrik,

Thanks for the update - good to hear it is working :D