Postby J.M. Finneran » Sun Jan 02, 2005 2:49 pm

I am looking to purpose a PC Scope and need to know which unit would be the best for my applications. I want to be able to measure voltage, current and the spike or spikes which occur during an electrical arc. I will be monitoring inrush currents, voltage spikes and the waveform the voltage spike causes. The testing has to do with the current spike a circuit breaker would see during an electrical event.

Thank you in advance for any assistance you can give me.
J.M. Finneran
 

Postby Sarah » Wed Jan 05, 2005 9:43 am

Hi

Thank you for your post.

What size voltages are you looking at? If the voltage is going to go above ±20V then you will need to use an attentuating probe to bring it into range.

Do you have a desired spectrum bandwidth, or resolution that the scope should be?

Best Regards
Sarah
 

Postby Carl Grotti » Thu Jan 06, 2005 5:02 am

Hi Sarah,

I can't be sure, but it seems as though the poster is testing breaker limitations. Maybe durations of high current when "trip" takes place? Like you mentioned, the voltage is a major consideration here. I'm wondering myself if mains are involved.

Carl
User avatar
Carl Grotti
Active User
Active User
 
Posts: 19
Joined: Thu Nov 21, 2002 5:17 am
Location: Missouri, USA

The same question

Postby Pavlik » Mon Jul 11, 2005 11:53 am

Signal frequency 10 - 30 MHZ
Signal magnitude 0 to 0,1V
Need FFT
randomly distributed events
Need FastFrame mode (Tektronix) or similar
Pavlik
 

Postby Sarah » Wed Jul 13, 2005 11:08 am

Hi

Thank you for the extra information.

With a bandwidth of 30MHz you would really need to use the PS3205 or PS3206 as these are the only scopes which would be able to see a 30MHz signal. Both of them have ±100mV ranges and above and so would be able to see a 0 to 0.1V signal.

I am not sure what FastFrame mode is, can you give me some more information as to how this works? I can then advise as to whether our products can do this. The software will certainly enable you to perform FFT.

Hope this helps

Best Regards
Sarah
 

FFT

Postby Pavlik » Mon Aug 15, 2005 10:24 am

There is more info. from the Tectronix user manual:

FastFrameTM
You can define and enable FastFrameTM (also called “segmented memoryâ€
Pavlik
 

Postby Sarah » Mon Aug 15, 2005 2:09 pm

Hi

Thank you for your post.

I do not believe that this is something that our units can do. I believe my colleague is responding to the email you sent with the pictures.

Best Regards
Sarah
 