Temperature & Current Logging.....

Forum for discussing PicoLog

Temperature & Current Logging.....

Postby GlennP » Fri Jul 19, 2013 9:14 am

Hi All,
I am trying to log two currents and a temperature from a PicoLog 1216 and PicoLog 1000 board. I followed the PicoLog 1000 manual for measuring the current (section 2.5) and have some sensible results from it, I now need to measure a temperature. I am attempting to use a K type thermocouple but seem to get odd values from it, 3.36mV and 4.27mV, for a room that is around 23'c I believe these values correspond to between 89'c to 104'c. I have followed section 2 Making Measurements 2.1 Measuring up to +2.5V and have connected R35 a 10K resistor to allow for the reading. This produces the above reading if AC volts are used instead of DC there is no value indicated. Should AC volts be used for this as the thermocouple will produce a negative voltage at low temperature? If AC volts are selected on the PicoLog software the value read is 0.0000? do I need to use a potential divider to get it to work, your advice would be most welcome.
Thank You
Glenn Patton.
GlennP
Newbie
 
Posts: 0
Joined: Tue Jul 16, 2013 2:20 pm

Re: Temperature & Current Logging.....

Postby Martyn » Fri Jul 19, 2013 9:22 am

A colleague will be contacting you through your email helpdesk ticket to explain how to use the logger to measure temperature. You can't plug thermocouples in without additional circuitry, cold junction compensation, and appropriate scaling of the values.
Martyn
Technical Specialist
Martyn
Site Admin
Site Admin
 
Posts: 2355
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots


Return to PicoLog

Who is online

Users browsing this forum: No registered users and 0 guests