Test and Measurement Forum

Pressure MEasurement using PT-104

Post any questions you may have about our current range of USB data loggers

Pressure MEasurement using PT-104

Postby junedfird » Thu Jan 03, 2013 11:45 am

I wish to measure pressure using PT-104.

Do I need to purchase 0-100mV output or 0-2.5V output pressure transducers?

How to connect the pressure transducers to screw terminal adaptors?

0-100mV transducers come in 4 wire configuration.

Posts: 0
Joined: Thu Jan 03, 2013 11:40 am

Re: Pressure MEasurement using PT-104

Postby Hitesh » Thu Jan 03, 2013 2:50 pm

Hi junedfird,

You can configure a channel on the USB PT-104 to be either in the range 0 - 115mV or 0 - 2.5V.

The user guide for the USB PT-104 shows how to connect wires to the device:


Please see section 2.6: Connection: PRT inputs. Only 3 of the wires need to be connected. Do you have any information on what the sensor wires correspond to?

You will need either a mini-DIN connector to attach to the wires or you can purchase a screw-terminal adaptor (PP660) http://www.picotech.com/document/pdf/DO172-PT-104_Screw_Terminal_Adapter_User_Guide-1.pdf.

You would need to scale the voltage readings to pressure in the PicoLog software.

I hope this helps.

Technical Specialist
Pico Technology
Site Admin
Site Admin
Posts: 2026
Joined: Tue May 31, 2011 3:43 pm
Location: St. Neots, Cambridgeshire