Tuning probes

Post any questions you may have about our current range of oscilloscopes
Post Reply
Mike_T
Newbie
Posts: 0
Joined: Thu Nov 29, 2012 7:57 am

Tuning probes

Post by Mike_T »

Hi -
The application note about tuning oscilloscope probes ( http://www.picotech.com/applications/ho ... robes.html ) says:
"The easiest way to compensate the LFC part of a probe is to input a square wave with a relatively slow edge but, importantly, no overshoot".

I just wanted to check - can I assume, firstly, that the signal generator of my Picoscope 2204 set to 1 KHz square and 2 V amplitude gives a signal suitable for tuning the PP787 (M1007) probes (with attenuation set at x10) and secondly, that there is no interaction between the generator and scope parts of the 2204 that undermine this operation. I don't have a separate signal generator and I have no independent way to verify the AWG output from the 2204.

I have not seen anywhere in the notes that says to tune probes this way - (probably because it is so obvious? :) )

Thanks, Mike

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Tuning probes

Post by Martyn »

That will be fine. You will even notice that when you start the signal generator from default the settings are 1kHz Square Wave ;)
Martyn
Technical Support Manager

Mike_T
Newbie
Posts: 0
Joined: Thu Nov 29, 2012 7:57 am

Re: Tuning probes

Post by Mike_T »

Thanks Martyn