Postby ridwan80 » Wed Oct 31, 2012 3:28 am

Hi, I am new user of Dr. DAQ and would like to do T measurement using already available K-type thermocople (not Dr DAQ T sensor). I would like to know, In which port should I connect the thermocouple ? What modification should I do in order to get the value (either in T or mV). Also, is it OK to measure the T up to 400 C using this devices ?


Thank you very much for your help.

Ridwan
ridwan80
Newbie
 
Posts: 0
Joined: Sun Oct 28, 2012 3:07 am

Re: Dr Daq T measurement using K thermocouple

Postby Hitesh » Wed Oct 31, 2012 4:26 pm

Hi Ridwan,

Unfortunately it is not possible to connect a thermocouple to the DrDAQ as it does not have cold junction compensation or the correct type of input for a thermocouple.

For 400 C measurements, you need some sort of suitable sensor that can connect via the External Sensor input and has an output voltage of 0 -> +2.5V. The device itself would not be suitable for placing in such an environment.

Hope this helps.
Hitesh

Technical Specialist
Pico Technology
Hitesh
Site Admin
Site Admin
 
Posts: 2058
Joined: Tue May 31, 2011 3:43 pm
Location: St. Neots, Cambridgeshire

Re: Dr Daq T measurement using K thermocouple

Postby ridwan80 » Thu Nov 01, 2012 2:56 am

Thank you for the explanation.
Now, I am using pico TC-08 for this purpose.

Best Regards,

Ridwan