Rise time of 2208 function generator

Post any questions you may have about our current range of oscilloscopes
Post Reply
jyrkipe
Newbie
Posts: 0
Joined: Fri Sep 14, 2012 8:55 am

Rise time of 2208 function generator

Post by jyrkipe »

I connected the output of 2208 function generator to the input using direct coaxial cable (good quality).

Function generator produced 1 MHz square wave.

Scope showed 200 ns rise time, which is far from the expectations for a 200 Mhz scope.

I suspect this is due to the function generator.

What is the rise time of the frunction generator in 2208?

Specs say the output impedance of the function generator is 600 Ohms. Is it possible, that it is much larger?

Regards
Attachments
pico.pdf
(91.6 KiB) Downloaded 406 times

Martyn
Site Admin
Site Admin
Posts: 4499
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Rise time of 2208 function generator

Post by Martyn »

Using an external generator the rise time can be seen in the low ns range.
Martyn
Technical Support Manager