High noise on a 4227

Post any questions you may have about our current range of oscilloscopes

High noise on a 4227

Postby kurt » Sat Aug 25, 2012 3:19 am

I've been testing my 4227 Picoscope with a pulse generator, and noticed a dramatic difference in the quality of the traces obtained at the 2 V range compared with the 5 V range (and higher). When you switch from 2V to 5V, the noise level goes up by at least a factor of ten, and my signals get an additional DC offset of -60 mV. In addition, the DC offset fluctuates by ~10 mV over the span of a few seconds. If I didn't know any better, I'd say that there's a floating ground in the scope.

Is this normal? I realize that it's to be expected that you take a noise hit when you go to higher ranges, but this is almost unusable.
Posts: 0
Joined: Fri Aug 24, 2012 12:04 am

Re: High noise on a 4227

Postby Martyn » Tue Sep 04, 2012 6:26 am

Are the readings you are getting within the specification 1% of full scale (20 °C to 40 °C) ? If not let us know.

The reason there is a change between 2V and 5V is due to the switching between different input circuitry, note the relay click as you change vertical scales.
Technical Specialist
Site Admin
Site Admin
Posts: 2314
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Return to USB PC Oscilloscopes