Picolog CM3 Current Data Logger slow response time

Post any questions you may have about our current range of USB data loggers

Picolog CM3 Current Data Logger slow response time

Postby Maxste » Fri Jun 22, 2012 11:42 am

Hi there,

I am using the CM3 data logger and it has such a slow response time that it is unusable at it is.
Basically if I input a square wave, the rising time and falling time are about several minutes whereas it is instantaneous (checked with an "classic" oscilloscope).

The conversion time (which I believe is what I am looking at here ) is meant to be 720ms, which is slow but fine for my application, not several minutes.

Any ideas?

Thanks
Maxste
Newbie
 
Posts: 0
Joined: Thu Jun 21, 2012 3:25 pm

Re: Picolog CM3 Current Data Logger slow response time

Postby Martyn » Wed Jun 27, 2012 7:49 am

It's possible that the voltages from the current clamp are at a very low level, a different clamp that provides a higher voltage for a given current should provide a faster response.
Martyn
Technical Specialist