Arbitrary waveform generator issue

Post your Linux discussions here
Post Reply
atulkumar
Newbie
Posts: 0
Joined: Fri Dec 23, 2011 3:13 am

Arbitrary waveform generator issue

Post by atulkumar »

Hello!

I am using a 2205 device and want to use the AWG to produce a wave. With a fixed waveform, it works fine as I call the ps2000_set_sig_gen_arbitrary function once with some values and the waveform is produced as expected. But if my program wants to change the data in the buffer (to change the waveform at runtime), I need to call the ps2000_set_sig_gen_arbitrary function again (because to the best of my knowledge the driver API does not provide any function to directly change the values in the device buffer that is used for AWG). The problem is that when the ps2000_set_sig_gen_arbitrary function is called, a random spike is generated by the device for a short time. If the ps2000_set_sig_gen_arbitrary function is called fast enough (even as slow as once every two seconds), the spike is clearly visible with the waveform. So the signal cannot be used for any meaningful purpose. Is there a way to get rid of the spike that is caused by a call to the ps2000_set_sig_gen_arbitrary. If not then is there a way to directly change the values in the device buffer?

Thanks,
Atul

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Arbitrary waveform generator issue

Post by Martyn »

When calling the ps2000_set_sig_gen_arbitrary function it stops and starts the scope which is why you are seeing the spikes. Unfortunately it is not possible to change the waveform on the fly due to the way in which the hardware operates.

If you wish to discuss this in more detail then please contact the support team at support@picotech.com
Martyn
Technical Support Manager

atulkumar
Newbie
Posts: 0
Joined: Fri Dec 23, 2011 3:13 am

Re: Arbitrary waveform generator issue

Post by atulkumar »

Thanks Martyn for your reply. I'll write to support@picotech.com.

But I dont think its the hardware limitation. Because when I use Picoscope 6 software's Signal Generator interface and then change some signal properties while the signal is being produced and viewed on a dedicated Oscilloscope, I dont see any spike in the waveform. Does Picoscope 6 also used the same API? I doubt.

Regards,
Atul

Martyn
Site Admin
Site Admin
Posts: 4491
Joined: Fri Jun 10, 2011 8:15 am
Location: St. Neots

Re: Arbitrary waveform generator issue

Post by Martyn »

Picoscope 6 does use the same API and I have seen this effect particularly when using a low signal generator frequency and then changing the wave type.