Triggering in XY mode ONLY when signal is low

Post any questions you may have about our current range of oscilloscopes

Triggering in XY mode ONLY when signal is low

Postby baries » Thu Nov 10, 2011 12:20 am

Hello ~ I'm using a PicoScope 4000 with software version 6 in XY mode (Lissajous figure - circle). I would like to run a several minute long test - and if no trigger occurs, and no signal is captured, then this test would pass.

:!: Is there a way to set the trigger to occur only when athe signal drops below a set amplitude? :!:

This is a circle test in a linear encoder application and the expected signal is 0.7vpp to 1.0vpp. If it drops below 0.7vpp, even for a moment, I would like to capture that point in the test.

Thanks!
Brian
baries
Newbie
 
Posts: 0
Joined: Tue Nov 08, 2011 10:40 pm

Re: Triggering in XY mode ONLY when signal is low

Postby Martyn » Mon Nov 14, 2011 10:56 am

If you open a Scope window, alongside the XY mode window, you can use this to set the trigger conditions.
Martyn