Just started to experiment with our new TH-03 module.
The high reference is being reported as 22044, when the manual indicates it should be within 500 counts of 43773.
Low reference is reported as 19986, well within the 500 counts of the base 20082.
Channels 1- 3 report between 21784 and 21852. This translates as 24-25 deg C in lookup table, which is about right by my desk thermometer.
If I use the given formula to correct these values with the high & low references, I get values between 40780 and 41562. This translates as below -30 deg C.
Is this a faulty unit, or am I missing something?
Picolog reads all three channels (currently around 25.5 deg C). Is there any way in Picolog to display the references?
hope someone can help.