An average 38mV p-p noise is measured by PicoScope 6 when set to +-5V. The p-p noise will decrease if the range is changed, but i need to measure the noise present on the output of a DC-DC supply which is in the range of 3-4.5v.
Reading the forum i see the mains from the laptop may inject noise into the oscilloscope, it could also come from RF captured by the probes. I have performed the following test:
Laptop on battery With the picoscope not connected to anything: 38mV p-p
Laptop on mains With the picoscope not connected to anything: 38mV p-p
Having the probe and ground connected to each other: 38mV p-p
Shorting signal and gnd with a wire at the picoscope bnc connector: 38mV p-p
Is this expected?
Is there a way i could measure more accurately my mV noise even with the 5V offset from ground?
PS. i have some saved waveform but see no method of uploading them atm.