Post subject: secondary wave form Posted: Wed Aug 15, 2007 8:18 pm |
|
 |
| Newbie |
Joined: Wed Aug 15, 2007 7:17 pm Posts: 1
|
|
Hi Dave, I just purchased a Picoscope 3423, I having problem with secondary king lead or plug lead set up, I have low reading or no reading at all with 20:1 attenuator installed. Is there way to fix it or wrong set up. please inform, James, thanks
|
|
| Top |
|
 |
|
davidj
|
Post subject: Posted: Thu Aug 16, 2007 12:00 pm |
|
 |
| Site Admin |
 |
Joined: Wed Oct 11, 2006 8:50 am Posts: 67 Location: St. Neots, Cambs, UK
|
|
Hi James, The 20:1 attenuator should not be inserted when using either the secondary clamp-on pick up or the COP probe. It would explain the very small signal you are seeing displayed !
Keep the 20:1 attenuator for primary ignition and injector work only. If you use the presets for these measurements you will see (20:1) on the left of the display in the units scale, indicating you need this attenuation for the scale reading to be true.
I hope this helps.