 Post subject: secondary wave form
PostPosted: Wed Aug 15, 2007 8:18 pm 
Offline
Newbie

Joined: Wed Aug 15, 2007 7:17 pm
Posts: 1
Hi Dave, I just purchased a Picoscope 3423, I having problem with secondary king lead or plug lead set up, I have low reading or no reading at all with 20:1 attenuator installed. Is there way to fix it or wrong set up. please inform, James, thanks


Top
 Profile  
Reply with quote  
 Post subject:
PostPosted: Thu Aug 16, 2007 12:00 pm 
Offline
Site Admin
Site Admin

Joined: Wed Oct 11, 2006 8:50 am
Posts: 67
Location: St. Neots, Cambs, UK
Hi James, The 20:1 attenuator should not be inserted when using either the secondary clamp-on pick up or the COP probe. It would explain the very small signal you are seeing displayed !
Keep the 20:1 attenuator for primary ignition and injector work only. If you use the presets for these measurements you will see (20:1) on the left of the display in the units scale, indicating you need this attenuation for the scale reading to be true.
I hope this helps.

_________________
Regards,
David


Top
 Profile  
Reply with quote  
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 2 posts ] 

All times are UTC


Who is online

Users browsing this forum: No registered users and 0 guests


You can post new topics in this forum
You can reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot post attachments in this forum

Search for:
Jump to:  
Powered by phpBB © 2000, 2002, 2005, 2007 phpBB Group

phpBB SEO