by ariaaudio » Fri Mar 30, 2007 6:45 pm
Thanks, Alan. I set the inputs to AC coupling, +/- 100mV sensitivity, and shorted the X1 probe together and nulled out Channel A. Viewing with Current/Filtered (to reduce noise), the trace set itself smack on the 0V line. I then connected the probe to a 9V battery and gave it a minute to settle. The trace rode a smidge higher, so that the dashed 0V line is visible below it. If I had to guess, it looks like maybe 1mV. Same with Channel B.
This looks like the coupling capacitors have some leakage. With 9V of DC this is hardly an issue, but going to 400VDC and using the X10 probe setting, I reckon the probe is presenting 40VDC to the scope, which appears to be less than the +/-100V maximum I see in the user's guide for this model. If blocked, of course, it would be zero volts, so all the ADC should see is any AC riding on the blocked DC.
But what I get with a 400VDC signal (AC coupling, X10 probe, +/-20V sensitivity) is a +8.2VDC offset on CH A, +7V on CH B.
With a external blocking capacitor (0.1uF/600V film) at the scope input, the leakage DC offset drops to +10mV. I reckon that my scheme of providing an external DC blocking capacitor would work, but since the 3206 has blocking capacitors inside, it might be possible to replace them with something that has less leakage?
We're quite capable of doing a simple capacitor replacement (he said, not having opened the 3206 to look within) -- is it possible to get a service manual or other documentation?
-- mike elliott