Externally triggering the ADC 212/3

Post any questions you may have about our data loggers and oscilloscopes that connect to the PC via the parallel port. This includes parallel port products connected to a PC using the USB parallel port.

Externally triggering the ADC 212/3

Postby Guest » Mon Oct 04, 2004 6:44 pm

Hello,

I cannot use a 1 volt, 20 msec, 50 kHz tone burst to externally trigger the ADC 212/3. What are the minimum requirements for the trigger pulse?
Guest
 

External trigger level

Postby Autonerdz » Tue Oct 05, 2004 12:21 am

The external trigger levels are fixed values. You can adjust for delay and slope but not level.

Positive slope external trigger level is about 3 volts.

Negative slope external trigger level is about 1.8 volts.

Your input will have to reach these levels to be compatible with external trigger. Hope this helps.
Tom Roberts
(The Picotologist)
http://www.autonerdz.com
skype: autonerdz
THE PicoScope Automotive Authority