Please recommend a PC Oscilloscope

 
Post new topic   Reply to topic    Pico Technology Forum Index -> Pre-Sales Advice
Author Message
jerrycho
Guest





PostPosted: Tue Jun 17, 2003 7:03 pm    Post subject: Please recommend a PC Oscilloscope Reply with quote

Hi,

I am a PhD student in Univ of Illinois at Urbana-Champaign, and currently working on energy saving for mobile devices. Basically, I change the CPU frequency/voltage to save energy.

Now I and my advisor want to buy a PC Oscilloscope to measure the variations of the energy consumption of a laptop or iPAQ. That is, when I change the CPU frequency/voltage, the Oscilloscope should show the changes of the input power of the laptop (e.g., the current or the voltage of the AC input). I expect that the power change is quite small.

Would you please recommend a desirable device for such a requirement? Thanks

Jerry
Back to top
markspencer
Site Admin
Site Admin


Joined: 07 May 2003
Posts: 610

PostPosted: Thu Jun 19, 2003 11:06 am    Post subject: Reply with quote

Hi,

The unit I would recommend is the ADC212-3. This unit has a 12 bit resolution and a range of voltage measurements , with the highest being ±20volts, and the miminum setting 10mV.

To measure the current you will also require a current clamp: 60A AC/DC and can measure from 10mA upto 60A. It also has two ratio settings 1mV to 10mA or 1mV to 100mA.

To measure voltage if it is above ±20 volts and below the low voltage directive (42.4V Peak or 30V RMS, 60 VDC) you will need a x1/x10 probe. If the voltage to be measured is above the low voltage directive you will need an isolating probe x10/x100, for safety reasons.

Best regards,


Best regards,
Back to top